Articles: surgery.
-
In addition to conventional C-arms which can be used for intraoperative imaging, C-arm image amplifiers with an option for three-dimensional imaging (ISO-C3D) are available to visualize reduction of fragments and positions of implants. In ten cadaver wrists intra-articular steps and intra-articular screw positions were simulated. Images obtained by conventional two-dimensional C-arm image amplifier, computed tomography and ISO-C3D were evaluated by four investigators using a questionnaire. ⋯ For ISO-C3D in fast 190° mode the rating was correct in 59%, wrong in 30% and uncertain in 11%. In the slow Iso-C3D mode the results were inferior with correct assessment in 51%, wrong results in 36% and uncertain evaluation in 13%. In our cadaveric study, ISO-C3D scans have been found valuable for intraoperative controls of implant positions and assessment of intra-articular steps.